deburca98
Hi all
I have a measuring device linked to a gaugeport and running a wedge software via rs232
I am having a problem with the software, basically the software worked fine for years i switched to a new (ish ) Pc dual core processor.
anyway every so often the following happens.
when taking a measurement sample and populating to excel the leading value is dropped and replaced with a zero, eg 8.76 is returned as 0.76.
this can happen once in every 250 entries or even 1 in every 50
baud rate is 9600 parity 1
i suspect it could be hyperthreading issue or a driver issue.
Another suspect is that FIFO is switched on
Can anyone help
forgot to say i am using a footpedal to enter data
could this be a framing error ?
Find more posts tagged with
Quick Links
All Categories
Recent Posts
Activity
Unanswered
Groups
Best Of
Advertisement
Advertisement
Comments
CreepingDeath
I don't use this software but...
If you're using "WinWedge" they say they provide unlimited technical support.
Link
I suspect it's a flow control issue, that it tries sending data too fast and loses some bits in the process.
This looks like it's a step in the right direction... auto hardware flow control setting in the last post here...
Link
So under Windows 7 go to
Start --> Control Panel --> Device Manager
In there expand the item for "Ports (COM & LPT)"
Select the COM port you use, eg. COM1, COM3 etc...
Right click on it and select "Properties".
In the "Port Settings" tab, try setting the "Flow control" option to "hardware" and see if it helps.
deburca98
os is windows XP
Disabled FIFO and enabled hardware on flow rate on Com port via device manager
Changed flow rate from none to hardware on winwedge software seems to have solved the issue.
Contacted the manufacturer winwedge no help at all discussed hyperthreading issues with the dual core processor
cheers will see if this helps